I’m having some difficulty using the QA480 with the QA402. I think it’s the FR measurement and export part. Since the software is slightly different from the 401. Any advise?
Thanks in advance!
I’m having some difficulty using the QA480 with the QA402. I think it’s the FR measurement and export part. Since the software is slightly different from the 401. Any advise?
Thanks in advance!
Hi @tonywood, which part specifically is giving you trouble? Can you share the export screen that shows the summary of the measurement?
The best I can get is about -116DB THD+N, Thought I should be getting closer to -120. I click freq response and it runs, then I export it. How to I configure it? Or is this correct?
Thank you
Hi @tonywood, from your measurement, your 2H is in the right place, and at -142 dbc we know that isn’t degrading anything. so, that means the noise is the issue. Can you add the measurement N-D dBV. That will show the noise while ignoring the distortion contribution.
We can roughly estimate what that number is. Your are at 19 dBV, and the THD says the noise is -115.88 below that, so your noise (absent harmonics) is likely about -96.9. That suggests whatever is out around 20 kHz isn’t a contributor even if not a harmonic (and likely it isn’t).
But from the N-D figure we can try to figure where to go next.