Maybe it is ghost chasing, but I tried to figure out the best “floor” readings I can get out from my QA403.
Connecting a QA403 to a tower PC running Debian Linux 13, Trixie, I was also able to run the QA40x software.
while connected via a straight USB-B to A cable, a loopback to itself in single-ended on left channel allow THD measurement to cross the -126 dB level and SNR is very close to 110 dB. All 3 other inputs have shorting plugs.
When using an external 5V wall adaptor to drive the QA403 via a modified USB cable, I am able to get a THD measurement under -127.5 dB and SNR almost hit 111 dB.
A direct USB cable show the voltage displayed by the QA403 to be 4.75VDC or a bit above.
The use of an external power supply and a modified USB cable show 4.97 VDC or a bit above.
The QA403 I have is able to go so low only if certain conditions are met.
The GEN1 output must be set to -0,23 dB / 1kHz.
Right channel is muted
GEN2 output is set to -150 dB / 23 kHz. (so as not to interfere)
FFT using flat-top window and set to 1024K.
Sample rate 48K
Full Scale Input is +6 dB
Average is 100
When 100 readings are completed, I read -127.51 THD or lower and 110.94 dB SNR or higher.
Changing the orientation of the unit, I was able to get the 60 Hz at lower than -142.7 dB.
The right channel does not have numbers as good. (2 or 3 dB difference on both SNR and THD)
Having the left channel set to -165 dB / 23 kHz and all 4 inputs shorted, the highest peak I get is for 60 Hz and it is blow - 155 dB. The rest of the 20-20,000 bandwidth is below -160 dB.