Hi there, I have just receied my QA403.

I was doing some loopback tests to get a grip with the analyser. Testing the AMP frequency response at about 96Khz I get the following plot. My quesiton is is the loss above 20kHz a feature of the analyser or is it my cable? I am using a 50 cm long BNC/BNC RG-58/U cable.

Thanks a lot Ivan

Hi Ivan,

Your plot is showing 192Ksps it looks like.

I took the QA403 on my desk and ran 3 sweeps, starting with 64K FFT for 48Ksps sample rate. And 96K I used a 128K FFT, and at 192K I used a 256K. For the sweeps, you can see the 48K has a bit of peak (0.01 dB at 20 kHz) and then drops off sharply. That is purely the digital filter inside the ADC.

For the 96 and 192K sample rates, you can see a bit of what looks like a first order filter rolloff before the sharp drop. That is an RC in the front-end of the analyzer. At 96K, you should be down about 0.1 dB at 37 kHz. And at 192k, the 0.1 dB point comes a bit sooner, but the 0.5 dB point comes a bit later.

The -3 dB point on my QA403 shows:

At 48K, ~22 kHz
At 96K, ~45 kHz
At 192K, ~88 kHz

That looks about the same as yours.

Thanks I re-run the tests and I get the same results as you. Thanks for the explanation. Ivan

1 Like

I cannot get the AMP Frequency Response (sine stimulus) to work above 20kHz.

When I type a value greater than 20kHz, I get an error. I was using a 192k sample/sec sampling rate.

I am not in mirror mode.

Screen shot below

Hi @DavidR. I think the error is because you also need to set at least 40Khz in the measurement options window, which is probably now set to less than 40Khz

1 Like

Hi @DavidR,

Yes, Claudio is correct. The error code hopefully is a reminder. Note the Measurement Stop Frequency is how you specify where the measurement will stop looking for harmonics. If you set the Generator Stop Frequency to 40k and your Measurement Stop Frequency to 50 kHz, so you won’t see any harmonics included in the measurement.

And so, if your aim is to measure the second harmonic of 40 kHz, then you probably want your Measurement Stop Frequency at 85 kHz, sample rate at 192k, and your Generator Stop Frequency to be 40 kHz.

Hi,

I’m looking here for my problem, may be a similar solution.

I want to measure the frequ response of my 500W valveamp (6x KT120) overall at ca. 70W into a 4 ohm load.

Mids 100% up, bass, treble and presence at 0:

Thats very ok for a bass amp.

With mids and presence at 0, bass and treble at full (its a fender tonestack) I get this:

Whats the “noise” in the highs? Without a signal I don’t hear anything in a conncted box, no high frequency oszillation in my scope visible.

Maybe due to the phase change caused by the eq? Exponential chirp methods assume equal phase delay for every frequency as they correlate the chirp in against chirp out under the assumption that frequency is a fixed function of time. A longer chirp will decrease the effect of this if so.

1 Like

Hi @cfortner, yes @MarkT is on to it. An exponential chirp spends as much time sweeping from 20 to 40 Hz (one octave) as it does 10k to 20k (also one octave). So, if we are sweeping at, say 1 octave per second, then the dwell time on each 1 Hz slice in the first octave is 1/20 = 0.05. And the dwell time on each 1 Hz slice in the last octave is 1/10k = 100u.

So, there is 0.05/100u = 500 times more energy in the first octave than the last octave. And so, you are seeing the uncertainty that comes from the much lower energy. You are at 192k, but you are only looking at the first 20 kHz. If you cut your sample rate to 48k, you’ll spend 4X more time at each frequency which should help with noise. And then increasing FFT will help too.

And unless you expect a big difference in EQ at mild versus hot levels of output, you might skip the attenuator and adjust the amp to give about 0 dBV of peak output. And then use 6 or 12 dBV full scale. And once you are satisfied with the shape of the curve at the lower level, crank it up and see how it changes, knowing that the added atten will increase the noise AND uncertainty.

2 Likes

Notice that in the first graph the dbV range is 120dB, while in the second graph it’s 35dB: I guess that the high frequency noise would be visible in the first graph too, if you put a 35dB range.

1 Like

I changed the range because the curve wasn’t good visible.

I‘ll try other settings, thanks!

With 48kHz and a FFT windows of 1024 I get this:

Fender Tone Stack

Mid 100%, Bass Treble and Presence 0%

Mid and Presence 0%, Bass and Treble
100%

Mid, Bass and Treble 100%, Presence 100%

Hi @cfortner, much improved!

So treble centered around 5 kHz, bass around 30 Hz, and mid centered around 250 Hz or so?

If you walked up to the amp and set all the controls at midpoint, would that give you a big scoop-out around 220 Hz or so???

Hi Matt,

sorry for delay, here the frequency response with all pots at 50% (but without presence, because the KT120 aren‘t installed).

The scoop is around 1kHz.

Apologies in advance for hijacking the thread, but I am searching in vain for a tutorial explaining how to generate graphs like this - that is a frequency response / bode plot with multiple sweeps / traces shown in different colours.

Or even an explanation of how to save a graph image with a white background instead of the default black.

Google is just throwing up links to MATLAB tutorials or useless AI gobbledygook.
I’d appreciate it if a human out there could provide a direction to what I am looking for.

Thanks in advance…

EDIT: this is in response to the first reply by Matt due to the graph he posted.

Hi. Click on the “Edit” menu, select “Setting” and in the window that opens under “Display Options” choose “Dark on Light”

OK, thanks.

Ignore the rest of my previous post. I’ve figured out that I need to run a frequency response test through the automated tests menu. Really intuitive… not.

Now…

Every time I add a new trace to a graph my previously entered Xmin/Xmax and Ymin/Ymax values are ignored and the graph is redrawn with full-scale settings. This is really frustrating and daft and tedious to correct. Is there a workaround or do I have to put up with this?

Alternatively, is there a way to open a fresh / blank graph for editing without have to first run something from the automated tests menu? I’d rather just do that and import my previously exported / saved bode plots.

Thanks…

Hi @12345qwerty, I think there are two work flows to think about:

  1. Interactively working on a DUT where you are tweaking knobs and looking for changes.
  2. Preparing a graph for documentation.

For the first, I generally use the following approach HERE. That is, a use right channel as reference and quickly iterate on the main screen. This can give several updates per second if needed (based on FFT) size, and then as I tweak the knobs on the DUT I can see in real time what is happening.

And then, if I need to document, I move to using the Automated Tests. When using the automated tests, just let the traces accumulate. For example, if you are wanting to look at tone knob behavior or frequency response behavior (as I showed above), just keep adding traces to the plot. Between runs, as you note, it will adjust to show the full range so don’t get hung up on prettying things up until you have collected all the traces. Once you have all the traces you want , then you go in an visually adjust everything to get the view you want and then export the graph.

1 Like