I’d like to be able to measure the maximum level and headroom of DUTs like EQ’s and compressors.
Obviously the QA403 isn’t able to do this as its max output is 20ish dB. So, I’m looking into solutions that can work in conjunction with the QA403 for balanced connections.
Hi @peppermint, the QA403 in balanced can do 16Vrms = 24 dBV = 26 dBu output. As you note, some pro audio can work up to 27.5 dBu = 18Vrms. I don’t know of an inexpensive way to get into that region. In a pinch, an OPA1612 powered by 4 9V batteries (+/-18V rails) at a gain of 2 could get you there. Run it inverting to avoid input common mode range issues, and you could get very close to the output rails which would give 12Vrms unbalanced or 24Vrms balanced = 30 dBu.
I’m not sure if I’m doing something wrong but I can’t get past ≈+20dBu out of the QA403 in balanced.
Edit: I just attempted to use both GEN1 and GEN2 with the same settings and max amplitude (20.22dBu). Is this the right approach? Either way, I don’t seem to be able to set a value higher than +18dbu in the automated tests in order to achieve something along these lines (AP image)…
It seems that I had the “Output Gain (dB) = +6dB” enabled in the “external gain” setting. It’s still a bit unclear to me what this does.
Also, I measured again today, with the same True RMS DMM and it indeed measures 996Hz. I really don;t know what happaned yesterday.
However, when I try to do some Automated tests like “AmpThd Vs OutLevel” there’s a higher limit and I get a warning that “value must be <=18”. This is regardless I choose dBV or dBu. This can['t be correct, right? Why the upper limit isn’t +26dBu when dBu unit is chosen?
Hi @peppermint, OK, I see the bug now. When you have an output gain value set, the max output level is still limited to 18 dBV no matter what. What should happen is the max value should be limited to 18 + user specified output gain. This will be fixed for the next release. For now, to get the higher levels out, you will need to set your output gain to 0 and remember the output level is +6.02 higher if you are in balanced mode.
Thanks very much for reporting this, and sorry I didn’t understand it right away!
I’ve downloaded and installed 1.216 hoping to have this issued fixed as stated on the log.
However, unless I’m doing something horribly wrong, the behaviour seems to be exactly the same.
The DUT is transformer balanced I/O and I have set the "Output Gain(dB)=+6.02dB in the “External Gain” window. When I set the GEN1= +4dbu my DMM reads 1.227 Vrms across the + and - so, I assume that’s all good.
However, when I try to do an automated test (AmpThd Vs OutLevel) I can’t go past +18dB exactly like I described in the past.
Thanks for the prompt reply. This solution seems to do the trick.
I will try to test it a bit further sometime next week but so far it looks fine! Thanks!
Could you please elaborate on where the 2.22 dBu value comes from? Wouldn’t it make more sense for the upper limit of the ‘Stop Level [dBu]’ to correspond to the maximum output level the QA403 can handle without clipping, regardless of the user’s selected output gain setting (e.g., 0 dBu, +6.02 dBu, or any other value)?
It’s helpful to think about your vantage point for talking about measurements. When you specify an input gain, you are telling the analyzer “there’s some input gain…but I’d like to talk about everything from the vantage point of the DUT”. And then, when you add in a 30 dB attenuator for measuring a large amp, you are telling the analyzer “I’d like to talk about the things from the DUT view point.” And so, if the analyzer measures 0 dBV, the real measurement (at the amp) is is 30 dBV. And the analyzer will indicate a 30 dBV signal in spite of just 0 dBV at the analyzer input.
And it’s similar on the output side. When you tell the analyzer there is 6 dB of output gain, you are telling the anlyzer you want the levels you specify to be measured at the DUT input. The actual level at the analyzer will be different, depending on the output gain you have indicated.