FOR THOSE JOINING LATE, THE DISCUSSION BELOW IS FOR A SPECIAL TEST MODE ONLY THAT WILL REMOVE ADC LIMITS. IT SHOULD ONLY BE USED BY ENGINEERS AND TECHNICIANS THAT UNDERSTAND THE SAFETY IMPLICATIONS AND SHOULD ONLY BE USED ON LOW-ENERGY CIRCUITS. THE NORMAL MAXIMUM INPUT TO THE QA402 IS +32 DBV AS WRITTEN ON THE TOP CASE AND FRONT PANEL.
Hi @ttako,
Let’s start with an example: When you apply 1V rms (no attenuator) to L+, that is converted by a fully differential amplifier (FDA) to two 0.5Vrms signals balanced that go into the ADC.
Similarly, if you apply two 0.5Vrms differential signals (no attenuator) to L+ and L-, that will be mostly untouched by the FDA and will gain result into two 0.5Vrms signals into the ADC.
So, the ADC doesn’t know if you are inputting an unbalanced (single-ended) or balanced (differential)
The internal rails on the QA402 are about 13.5V, and the front-end passive attenuator is 24 dB. So if you input a single-ended 24 dBV signal with the attenuator active, it will be reduced to 0 dBV by the attenuator, and the FDA will convert it to two 0.5Vrms signals (balanced) just as above.
In the test mode, if you input a 42 dBV signal into the L+ alone (with L- grounded) with the attenuator active, the attenuator will knock that down to 42-24 = 18 dBV = 8Vrms. The tips of this signal will be around 11.2Vp, and this is within the common mode range of the opamps with the +/-13V rails. That 8Vrms signal will be attenuated by an active attenuator to 0 dBV for the ADC.
So, the short answer to your question is: In the special test mode, you can apply up to +42 dBV to either the L+ input (with L- shorted) OR the L- input (with L+ shorted) OR a fully differential signal of 36 dBV = 63Vrms to L+ and L-. All of these will be reported as a +42 dBV signal.
If you are not in the special test mode (and thus +32 dBV max), then the maximum input to L+ is +32 dBV (L- shorted) OR the L- input (with L+ shorted) OR a fully differential signal of +26 dBV = 20Vrms to L+ and L-. All of these will be reported as a +32 dBV signal.
Note that the QA402 overload characteristics (in terms of signal quality) will be similar to applying 0 dBV to the QA402 with the 0 dBV full scale mode selected. In other words, when you are at the limit, clipping may be happening.
Finally, remember to limit the time at the higher AC voltages. The longer you subject to the QA402 to the special test mode voltages, you increase the risk of the front-end aluminum caps degrading. This will manifest as the low-end response of the QA402 increasing.
If you need to make repeated measurements at very high voltages an external attenuator is your best choice.
Let me know if you have more questions!