Hello! I’m using a QA401 with Analyzer version 1.923. We are using the QA401 to automate effects pedal testing through Tractor - we have been using it for go/no-go tests using amplitude measurements for a while, but I want to add masks to the mix and am having trouble.
When using the FreqResponseChirp plugin, the result is offset by almost exactly +6dB from the Freq Response measurement done by the analyzer. The chirp amplitudes are both -30dB. I tried setting the external gain field to +6dB to nudge it back, but it also offsets the measurement that the QA401 is actively taking. Is this a bug, or is it user error on my part?
The 6 dB is a good clue. A few things that might be impacting your result: You have an input offset specified in the DBV menu, or the pre-analyzer input gain is non-zero in the Tractor settings, or you are making a differential measurement when single ended is expected (or vice versa).
Thanks, that test did work. Based on this, I don’t think it’s the act of reading the mask file back into QuantAsylum (or Tractor) that is causing the issue. It appears that some setting for the chirp plugin is introducing the 6dB offset. I had a similar thought as you that maybe I had a gain error somewhere.
I don’t think this problem was caused by my Tractor settings. I recorded traces for masks with the attenuator off, and they work perfectly whether I set the input level to 6dBV or 26dBV in Tractor. I think the attenuator setting is lost when generating the mask data. When I run the “AmpFreqResponseChirp.dll” plugin, it seems to compensate for the attenuator setting in a different manner from the Analyzer software. If I save traces at +6dBV, generate a mask in the graph tool, and import that back into Analyzer, it works. If I do this at +26dBV, the +6V offset appears.